From: Claude Cantin (cantinc@uhnres.utoronto.ca)
Date: Wed May 23 2001 - 17:06:55 EST
Hi All
I have been unable to detect a PharRed signal from samples I have been
given, and don't know if the problem is instrument or labeling error.
MoFlo set up is as follows:
Hammatsu 957-08 PMT all around
Omega Optical Filter Set XF501 (consists of 710 DRLP,
670DF40, and 740EFLP)
Melles Griot 35mW HeNe
Red laser alignment seems fine: "Molecular Probes AlignPlus beads for
633" are almost off scale if no ND filter, at 399V, linear gain 1. BD's
APC like Calibrites show excellent separation + / - ( >2 logs) 580V.
Whether I measure APC in FL5 (remove 710LP dichroic, use 670/40BP) or
FL6 (710LP plus 670BP), I obtain good APC signal. But no PharRed with
the 710LP dichroic / 740EFLP in FL5, and no PharRed signal even if I
remove the 710LP dichroic.
BD Pharmingen Technical data sheet recommends a Hammatsu R3896 PMT for
improved red sensitivity. Is the Hammatsu 957-08 deficient in the >740
range?
If not, what's up?
Thanks,
Claude