Re: Fluorescent zinc chelators

From: Bill Telford (telfordw@box-t.nih.gov)
Date: Mon May 21 2001 - 12:08:48 EST





Hi Kathleen...

Fluorescent calcium chelators can be used to measure cations other than calcium, but controlling for calcium and magnesium binding makes this difficult, particularly for low-concentration trace metals like zinc.  If you're courageous ;-), you might want to try a compound called "Zinquin", a somewhat cell-permeable fluorescent zinc chelator that is commercially available (Dojindo Laboratories in Japan and Toronto Research Chemicals in Canada are two suppliers I'm aware of).  Zinquin is an ethyl ester -modified form of its parent compound N-(6-methoxy-8-quinolyl)-p-toluenesulfonamide (TSQ), a non-permeable zinc chelator that can be excited with UV or violet light (Hg lamp, Ar-ion UV or Kr-ion UV or violet lines all work) and has a broad emission peak from about 460 to 510 nm.  Long ago, I used TSQ to do fluorometric zinc measurements in a cuvette and to label fixed/permeablized cells for zinc for microscopy.  In cuvette experiments, TSQ  fluoresced brightly only in the presence of zinc - cadmium and silver (similar atomic radii to zinc) showed some fluorescence, but these metals are not normally found in cells.  Calcium, magnesium, copper and iron did not cause the chelator to fluoresce.  We did some unpublished fluorescence microscopy work with it to look for zinc localization in fixed lymphocytes - TSQ precipitated and fluoresced when bound to zinc at physiological pH, showing sites with high zinc concentrations.  I've tried Zinquin more recently out of curiosity to see if it would label live cells for both microscopy and flow - with it, we can roughly monitor zinc uptake in lymphocytes incubated with micromolar zinc concentrations.  However, I haven't done specificity tests on this more permable reagent, nor established concentration standards.  A few labs have used this compound to measure intracellular zinc levels in lymphocytes and in spermatocytes by flow (refs), but no one (to my knowledge) has done a rigorous job of standardizing this reagent as has been done for indo-1, fura-2, etc.  There are potential drawbacks - it is not clear if these chelators bind to free zinc only, or can also detect zinc bound to metal binding proteins like metallothionein.  TSQ and Zinquin may also bind to (although not fluoresce with) other trace metals to some degree, which may cause an underestimation of zinc levels when other trace metal concentrations are high.  Nevertheless, you might want to give this reagent a try, with the caveat that it has not been as well-characterized as other fluorescent chelators.  A good opportunity for original technique development, though!  I can send you some data and more info if you are interested.

Regards,

Bill Telford
CCR-NCI-NIH

Some refs...

Andrews et al., Cytometry 21, 153 (1995)
Zalewski et al., J Histochem Cytochem 42, 877 (1994)
Coyle et al., Biochem J 303, 781 (1994)
Zalewski et al., Chem Biol 1, 153 (1994)

At 12:39 PM 5/17/01 -0500, Kathleen Schell wrote:
>
>I have an investigator who wants to measure levels of Zn++ in his
>cell lines.  Has anyone had any experience with Fura-2 that could
>give us some pointers?  We are planning to excite this dye with a
>multi-line UV and collect the emission at 510.  The literature that
>we have found just uses spectrophotometry to measure this metal.  Can
>we do this?
>Thanks,
>Kathy
>--
>--------------------------------------------------------------
>Kathy Schell
>Supervisor, UWCCC Flow Cytometry Facility
>600 Highland Ave.  K4/535
>Madison, WI 53792
>Voice:  608-263-0313
>e-mail:  kschell@facstaff.wisc.edu



This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:40:20 EST