Re: DAPI vs PI for dead cell discrimination

From: Eric Van Buren (aa9080@wayne.edu)
Date: Mon Feb 26 2001 - 11:53:32 EST


Adrian Smith,

We've had success with Calcofluor white (CFW) as a UV-excited live-dead dye. If
you just compare %-pos with PI, CFW yields higher %-pos. On closer examination,
you will see that CFW also appears to label "cell fragments" that PI does not
(because these cell fragments possibly do not contain DNA). One reference from
1994 lists these sources: American Cyanamid, Difco, Polysciences, Remel, and
Sigma (http://www.nejm.org/content/1994/0331/0021/1459a.asp). Interestingly,
this reference also lists another source for CFW, "... Liquid Tide with Bleach
Alternative, but any product containing a nonchlorine 'optical bleach' should
do as well."

For more information, consult Howard Shapiro's "Practical Flow Cytometry". He
also provides a more useful reference for using CFW with animal cells.

Eric


>Hi all,
>	We are currently evaluating a B-D LSR in order to decide
>whether it meets our requirements. One possible configuration we are
>considering would use DAPI excited by the UV laser (325nm) for
>live-dead discrimination in place of PI.  I know we don't need to
>devote a channel to PI but some users would prefer to have a separate
>live/dead channel and in most experiment we wouldn't be using the UV
>laser for anything else.
>
>My question is has anyone used DAPI for live-dead discrimination? How
>does it compare to PI? Are there any particular issues we need to be
>aware of?
>
>
>Thanks,
>
>Adrian Smith