Re: DAPI vs PI for dead cell discrimination

From: Howard T Petrie (h-petrie@ski.mskcc.org)
Date: Fri Feb 23 2001 - 18:30:08 EST


We've been using DAPI for dead cell exclusion for at least the last
five years on a Vantage, and for the last year or more on the LSR.
It is absolutely identical to PI in terms of the specificity of
staining/non-staining.  For viability, we use DAPI at 100
nanograms/ml when using cells at ~5 million/ml.  If you go up in cell
density, you will probably need a higher concentration of DAPI,
depending on how many dead cells you have.  Go for it.

>Hi all,
>	We are currently evaluating a B-D LSR in order to decide
>whether it meets our requirements. One possible configuration we are
>considering would use DAPI excited by the UV laser (325nm) for
>live-dead discrimination in place of PI.  I know we don't need to
>devote a channel to PI but some users would prefer to have a separate
>live/dead channel and in most experiment we wouldn't be using the UV
>laser for anything else.
>
>My question is has anyone used DAPI for live-dead discrimination? How
>does it compare to PI? Are there any particular issues we need to be
>aware of?
>
>
>Thanks,
>
>Adrian Smith

--