Re:

From: Howard Shapiro <hms@shapirolab.com>
Date: Thu Feb 24 2000 - 20:35:29 EST
Chris writes-

>I've got to drag my UV laser out of semi-retirement and I am wondering
>just what the
>latest word on alignment beads is.  I need to align both my 488 and 325
>UV, lasers.

Before aligning a 325 nm (I assume helium-cadmium) laser which hasn't been
used for a while, you may want to measure the noise on the laser.  Leaving
a He-Cd idle for a few months almost always results in helium build-up in
the tube, which will give you really high noise levels (power output
fluctuations of greater than 50%).  Since the fluctuations in laser power
are not compensated for by the electronics in any commercial flow
cytometer, they will translate into atrocious CV's no matter how well your
laser is aligned.  This not only makes it impossible to align the
laser, it
makes it impractical to use it (even ratiometric calcium measurements get
ratty when the noise is really high, and you can just plain forget
Hoechst,
DAPI, etc.).

Polysciences 2 um yellow-green beads, which are convenient alignment
standards for 488 nm excitation, also excite at 325 nm; they're not great,
but they probably will let you know if your CV is 25% as opposed to 2.5%.

-Howard
Received on Fri Feb 25 12:47:59 2000