From: Marty Bigos (bigos@stanford.edu)
Date: Fri Dec 03 1999 - 14:19:39 EST


We have had users getting recoveries significantly different from the 
Facstar counter as well. Although, at times, this could depend on the 
sample prep and sort conditions, a test run in parallel between a 
group that got consistently good agreement and a group which had low 
(< 80%) agreement revealed problems with the method of handling and 
counting the cells in the collection tube. Coating the collection 
tube with serum to prevent cells sticking, spinning the cells down 
harder, and using the appropriate cover slip in the hemocytometer 
resulted in the counts of the collection tube being within 10% of the 
counts on our Facstar.

Marty Bigos
Stanford Shared FACS Facility