4 colours on FACS Calibur

From: Michael Ormerod (Michael_Ormerod@compuserve.com)
Date: Sun Nov 28 1999 - 15:30:38 EST


Ihave not worked on this system myself but I am told that the He-Ne laser
will excite the Cy5 on the PE-Cy5 conjugate directly, which accounts for
the need to apply so much compensation. Looking at the excitation/emission
spectra, this is what would be expected.

I think that it might be best to use PerCP as the third dye.

Michael Ormerod