RE: Cyto Chex Summary

From: Gerhard Nebe-von-Caron (Gerhard.Nebe-von-Caron@Unilever.com)
Date: Thu Oct 07 1999 - 05:55:23 EST


Just a question regarding this reagent. I gave up on it as I wanted to use it
for preservation of cell function. However it was not compatible with heparin
blood which is normally used for functional cell assays and also it was said to
fix the cells. Thus I wonder how the product actually performs on lysing the
cells later on for example with FACS-lyse or equivalent. Remembering the
differences in scatter patterns one gets from cells lysed with different
solutions and then measured in different instruments I am not too surprised
that some people might find the cells looking strange in some instruments.

Regards
Gerhard


-----Original Message-----
From:	Abby Kelliher [SMTP:allena@helix.mgh.harvard.edu]
Sent:	Tuesday, October 05, 1999 5:34 PM
To:	Cytometry Mailing List
Subject:	Cyto Chex Summary


I have combined all of the responses to my Cyto Chex queary and have
included those responses (with names removed) below.  In summary, most
people responded positively but there were two people who have had poor
results with this product.  There seems to be conflicting results regarding
use of this product with whole blood.  

My own testing, while progressing slowly is encouraging.  I have tried it
on a positive B-ALL bone marrow with no differences one week later.  I have
also tried it on a CSF specimen and the specimen held in Cytochex looked
better than the one we processed fresh.  

I thank everyone for their detailed experiences with the product and hope
that this turns out to be the answer to my weekend prayers!!  I am also
interested in further experiences with this reagent.

Abby Kelliher
Clinical Flow cytometry Lab
Mass. General Hospital
Boston, MA  02114


Positives:
----------------------------------------------------------------------------
--------------------------
We have used Cyto Chex over weekends on blood and it works well.
----------------------------------------------------------------------------
--------------------------
I usually don't keep blood or bone marrow specimens for a week before
flow analysis (usually it's 2-3 days, due to weekends and my being off
one day for working the weekend in the coag section of our lab), but the
vast majority of the time, I get good results with keeping blood/bone
marrow samples in Cyto-Chex--as long as we get the specimen into the
Cyto-Chex the same day that it has been obtained.  I get VERY good
results with keeping tissues in Cyto-Chex, as long as I disaggregate the
tissue the same day that it has been received by the pathologist
assistants.  I think the occasional junky-looking specimens have more to
do with the specimen itself, not the Cyto-Chex, per se.
----------------------------------------------------------------------------
--------------------------
We have been using Cytochex for a couple of months now and are pretty happy
with it.  For the most part we get very good light scatter and Ag
expression.  We use it on late Friday afternoon specimens and use it only on
non-bloody specimens.  I would not recommend it for whole blood since we
have had some problems with it too (as one person described it today on this
list).  If you want to do bone marrows or bloody specimens, you need to lyse
it first and then store it in Cytochex.
The only problems we have encountered so far, are:
1.  DNA ploidy is often near-diploid even on benign samples after Cytochex
storage.
2.  CD 10 expression seems to be often non-specifically positive.  But if
you compare against your internal negatives, you are able to determine
whether the positivity is true or non-specific.
3.  Problems with peripheral blood as described.
We really try to only use it when it is too late on Friday to analyze but it
is not a stat specimen.  Over all, it is not the perfect solution but pretty
close!
----------------------------------------------------------------------------
--------------------------

We have been using the cyto chex for over a year, but our validation shows a
good correlation up to 4 days, we used over weekends or holidays. Do not
check viability by PI or 7AAD after cyto chex some permeabilization is
present.
----------------------------------------------------------------------------
--------------------------
I have carried out some stability work on cytochex, but it was limited to 
whole blood and ORTHO TRIO reagent i.e CD3/4/8/16/19. We tested over an eight 
day period at room temp and 4degC. and the results were very good. If I 
remember rightly, the manufacturers state that the sample should be viable up 
to 7 days at 4 deg C for both transport and storage. The only issue to keep 
in mind is obtaining accurate dilutions for absolute counts.
----------------------------------------------------------------------------
--------------------------
I also have had excellent experience with this product.  Although I was one 
of the original "testers" of this product I have taken it further and used it 
in my lab for my own purposes.  For example, I stimulated normal peripheral 
blood with CD2/2R to obtain high expression of CD69.  When the blood is put 
in Cyto-Chex the specimen can be used as a positive control In our activation 
assays for at least one week with no loss of antigen expression.  We're 
pursuing other uses for this great product.
Also , when we need a "normal" control (fresh blood) we draw only on Monday, 
put the blood in Cyto-Chex and use it the rest of the week so we don't have 
to draw someone daily.
I hope you keep experimenting with other uses for this great product.  I have 
nothing but good to say about it.
----------------------------------------------------------------------------
--------------------------
 We have looked very extensively to the Streck reagents:
-Platelet Chex (unfortunately discontinued)
-CD Chex
-CD Chex Plus
-Cyto Chex

This Streck company has some great products. We evaluated their controls and
stabilizers for our quantitative flow cytometry application kits. You must
know that quantitative flow is not very forgiven and we found great
performances for these products on various measurements.
It did not work for everything. 
----------------------------------------------------------------------------
--------------------------
Negatives:
----------------------------------------------------------------------------
--------------------------
I tried it out in my lab in Massachusetts to permit fedexing samples from
the South.  A single sample of human blood (mine) was treated with CytoChex
on day zero and then analyzed daily for the next 4 days.  The results of
CD4+ and CD8+ were significantly different on day 1 and subsequent days.  As
a result, despite our great hopes according to  claims made, we did not
adopt its use.
----------------------------------------------------------------------------
--------------------------
I hate to comment because I haven't finished my own evaluations. But, it's not
at the top of my list of to do's because I wasn't overly pleased with the
initial comparison data. 

I did have better results for subsets than the one who got varying subset
values
on days 2, 3, 4. We use a CD45 v. Side gating strategy for subsets and got
comparable percentages on days 2 and 3 - I didn't test beyond that. The
scatter,
however, looks pretty bad as grans don't seem to hold up well. While it's true
that that is of no great consequence with my particular gating strategy, it
could lead to a very messy specimen for those using light scatter gating,
since
the grans frequently fall into the gated area. And it just looks bad - I hate
reporting out stuff that is so obviously aged, even if it has been
validated on
some other bunch of samples.

Also, we tried Cytochex on some lymph node and bone marrow specimens. Same
story
- ratty looking cells. Worse, though, was that it appeared that the surface
light chain expression was undemonstrable after Cytochex storage of the single
cell specimen preparations. If I can't demonstrate clonality if present, I
might
as well not run the specimen.

I knew that there was some literature out there using this preservation
method,
so asked my new Streck salesman for it. He zipped it right over to me - you
should ask them for it via fax - or I'll send it to you if you want. What the
paper _apparently_ says (ok, I haven't read it all the way through yet,
myself,
but a co-worker reported the findings to me) is that the reagent would keep
the
neutrophils from upregulating or otherwise changing the CD11b status of the
specimen. The pictures I did look at(!). It looked to me like a large part of
the specimen had turned to debris after three or however many days. While
it may
be so that the percentage of CD11b on the grans was comparable to that of the
original test, I don't see how the measurement is even valid after losing a
large part of the population.

So that's my take on it. I haven't given up, but really don't see using it. I
purchased a ton of it, too, expecting that it would be just what I needed (the
lab's closed on Sat/Sun).