RE: Autofluorescence in bone marrow samples

From: Gerstein, Rachel (Rachel.Gerstein@umassmed.edu)
Date: Tue Oct 05 1999 - 10:59:10 EST


Dear Andy,

Human or mouse ?  Do they have a "spare" channel for a dump gate ? Can they
use a myeloid marker to gate these cells out ?  I have used anti-GR-1 in the
mouse, which is usually very good as a negative gate for the large, high
autoflourescent cells of myeloid/gran lineage.  Also, what about a viability
stain like PI - those autofl. cells may be stain slightly with PI, even when
viable.

Good luck,
Rachel

=======================================================
Rachel M. Gerstein, Ph.D.
Department of Molecular Genetics and Microbiology
Graduate Program in Immunology/Virology
University of Massachusetts Medical School
55 Lake Avenue North
Worcester, MA 01655-0002
(508) 856-1044
(508) 856-5920 (FAX)


> ----------
> From: 	Andy Lane
> Sent: 	Tuesday, October 5, 1999 7:48 AM
> To: 	Cytometry Mailing List
> Subject: 	Re: Autofluorescence in bone marrow samples
> 
> 
> Dear All,  
> 
> I'm posting this for a colleague, so don't have all the details, but hope
> that someone may have some suggestions.
> 
> They are using "whole blood" bone marrow preparations in analysis of
> various
> leukaemias/lymphomas, and consistently find a problem with autofluorescent
> cells that give signals in both the FL1 and FL2 channels (they have that
> typical non-specific diagonal look on a dot-plot of FL1-FL2). The
> non-specific signals appear even in the absence of any antibody, so this
> seems to be a true autofluorescence rather than non-specific antibody
> binding.
> 
> They are reluctant to gate out on FSC/SSC after back-gating as they feel
> that they risk losing important cells.  Does anyone have a simple way of
> overcoming this problem, or even know for certain what the cells are
> likely
> to be.  They are not seen after ficoll separation, and scatter suggests
> that
> they are of myeloid origin
> 
> 
> Andy