If the system is to maintained sterile then you need not change filters and
chance contaminating the system.
The large Pall filter can be used with  >20 liters of sheath without having
to be changed.
I use a (PALL)Gelman Versacap P/N 12130 from Fisher that does the same
thing.  It is a bit cheaper and is actually a double .4u with a .2u output.
This is a clear capsule and comes marked as "Sterile".
I use this with one tank of sheath (7 liters).  Because I sort cells for
patient use I dispose of the filter after the 7 liter usage.

jim