You need to either pre-soak or wash the blood before adding the antibody or
it is aborbed by elements in the plasma. We wash the blood three times with
PBS + 1%FBS and then stain and they come out great. You should be able to
test it on normal whole blood to make sure it is working right and you
should get a 50/50 split on CD19 cells.

Brian Newsom
Director, Flow Cytometry
Center for Cell and Gene Therapy
Baylor College of Medicine

-----Original Message-----
From: Sebastian, J, Gelderbloem, Mr [mailto:sjg@samiot.uct.ac.za]
Sent: Wednesday, July 14, 1999 10:10 AM
To: cyto-inbox
Subject: Re: CLL PHENOTYPE



Dear Flowers

I have a slight problem and I am not sure what the solution is.

We use polyclonal Kappa and Lambda from BD for our chronic 
leukaemia panel. We have been setting the samples up using whole 
blood and for some or other reason the Light Chains are coming up 
negative everytime. We only bitmap lymphoid cells.

Is there anyone with some advice.

Thanks 
Sebastian