Actually, I was able to see E. coli and Staph aureus rather well on my
Coulter XL-MCL. I just used a bead set to set the scatter to between
0.5-5 um, and then used viably stained (BIO-RAD) and heat-killed + PI
stained basteria to set the live/dead gates.
Some data from those studies are included in the recent publication by
Goodson et al. 1999. Antimicrob Agents Chemother. Jun;43(6):1429-34
Marty Giedlin
______________________________ Reply Separator _________________________________
Subject: Re: E-coli on Coulter Epics XL
Author: adrian rubio <genius@nmt.edu> at SMTP
Date: 7/8/99 2:59 PM
I don't think so... they are too small unless you can immunologically bind
them to some sort of resin bead because Ecoli are much smaller than
RBC's and would fade in with the noise.
_Adrian