Apoptosis

From: DARZYNKIEWICZ ZBIGNIEW (DARZYNK@nymc.edu)
Date: Fri Feb 19 1999 - 11:18:04 EST


The "best" positive control:
HL-60 cells treated with >0.15 uM captothecin (CPT) provide a reliable model
of apoptosis. Most apoptotc changes (mitochondrial, plasma membrane, DNA
fragmentation, nuclear fragmentation) occur during the initial 4 h of the
treatment. The advantage of this model is that only S phase cells undergo
apoptosis (Del Bino et al., Cancer Res., 51: 1165, 1991). Thus, G1 and G2/M
cells, within the same sample may serve as a negative control. The critical
point is the cells have to rapidly progress through S phase to be sensitive
to CPT. It is a collision between the progressing DNA replication fork and
the lesion iduced by CPT that provides the signal inducing apoptosis. Any
slowdown in S phase progression, therefore, such as  due to higher density
of cells in the culture (subconfluency; > 800.000 cells per ml) makes them
less sensitive to CPT.
 
2. The issue as to whether the second "p" is silent is a subject of long and
ongoing dispute. Interestingly, it become apparent quite recently that the
term was already used by the father of medicine Hippocrates, to describe the
falling of the bone fragments during healing , i.e. in the context related
to its common use (see Esposti: Cell Death & Differ.; 5: 719, 1998). As a
Greek word, it should be pronouced with two "p" (see Funder, Nature, vol .
371, 1994 (Sept. 8 issue)"Apoptosis: two p or not two p". English authors,
however, often transform its prononciation to English language suppressing
the second "p". 
Zbigniew Darzynkiewicz