Eric and Arnold does a wonderful job of explaining the reason and
solution for your problem, but perhaps I can still add something if the
cleaning of the impeller and its three metal ends (magnets) does not help.
I had the same problem with my Innova 90 and the cleaning of the impeller
did not help. I recommend you to buy three small magnets and drill them
in to those 3 empty heads of the impeller. Be careful and check that you put
those magnets into the impeller in the same direction like the others are.
In this way you lure the power supply box to sense a doubled flow rate.
Coil
OOOOOOOOOOOOO....................
OiiiiiiiiiiiO............Pulse out
IIIIIIIIIIOOOOOOOOOOOOOIIIIIIIIII
I oMo I
I oMo I
I oMo I
I ooo I
I oo ooo ooo I
I ooo ooo ooo I
I ooo ooo ooo I
I ooooooooo I
I oXo I
I ooooooooo I
I oMo ooo oMo I
I oMo ooo oMo I
I oMo ooo oMo I
I ooo I
IIIIIIII ooo IIIIIIII
I I I ooo I I I
I I I ooo I I I
I IIIII IIIII I
I I IIIIIIIIIIIIIIIIIIIIIIIIIIIIIIIII I I
I I I I
I I I I
I I I I
I I I I
IIII IIII
Water out Water in
*Letter M represents magnet.
Best wishes,
Mika
Mika Korkeamaki
University of Turku
Finland
On Tue, 26 Jan 1999, PETER BERCZ wrote:
>
> ** Reply Requested When Convenient **
>
> To anyone who may have come up against this problem: On my Coulter
> EPICS Elite I have an INNOVA 305 Argon Ion laser, only about 100
> hours on it. The water flow to the power supply via a new filter is
> about 3.5 gal/min (verified by measuring flow rate by volumetry). In and
> out flow to and from the power supply is correctly hooked up, I know it
> can cause problems if it is hooked up in reverse. The filter is not clogged,
> the hose is not crimped. Still the remote control display registers 0
> flow, consequently the laser cannot be powered up. Did anyone out
> there encounter this problem? While I am in touch with Coulter, maybe
> someone knows, Is there a valve or a flow activated switch that got
> stuck? Thanks for any advice. Peter Bercz, USEPA/NERL EERD
> Cincinnati
>