>We are looking for working reliable methods for DNA content staining for
>alive cells. I know that few of them are exist but I heard that people
>always have problems with them.
>
>We'll appreciate any references, advices.
>
>Thank you.   Marina

Greetings Maria,

        I have recently tried 'Syto16' from molecular probes, it's a
cell-permeable Nucleic acid stain that is green fluorescent, the big
problem with it is that it has some affinity for RNA and, in some cell
types may also stain cytosolic components, I know this sounds as though
Syto16 is totaly unsuitable for DNA analysis but, it all depends on what
you want to do....       Differentiation of G0/G1, S, G2/M would be
difficult using Syto16 alone, as the cell cycle peaks are convolved with
the RNA/cytosolic distribution. If you simply want to analyse/seperate
cycling/noncycling cells however, combining Syto16 with forward scatter
gives a good correlation when sorted cells are fixed,restained with
Propidium and reanalysed.

best wishes

                Arnold.

--------------------------------