HCFA - Sample letter

From: Richard McFarland (mcfarland.richard@pathology.swmed.edu)
Date: Mon Nov 09 1998 - 17:15:02 EST


>
>        The deadline for receipt of comments on the proposed regulations
>is 11/11/98.  This means that letters must be sent by overnight mail by
>noon 11/10!!!  Below is a copy of a letter that is being sent from a
>pathologist at our institution.  This may provide you with helpful
>information and concrete arguments to use.  If you absolutely can't mail
>your letter by noon tomorrow, the FAX number is 202-690-5863.  However,
>this fax line may be jammed on 11/10, making it impossible for your
>comments to be heard - so rely on it only as a last resort!
>
>SAMPLE LETTER
>Judy Ballard and Nancy Rubenstein
>Facilitators, Negotiated Rulemaking Committee
>DHHS Departmental Appeals Board
>Hubert H. Humphrey Building, Room 637D
>200 Independence Avenue, S.W.
>Washington, D.C.  20201
>
>Dear Ms. Ballard and Ms. Rubenstein:
>
>        As a pathologist practicing flow cytometry, I am writing to
>express my concerns regarding the "Recommended Medicare National Coverage
>Policy for Flow Cytometry".
>        For clarity, I have reprinted the document text below, and
>inserted my comments in the areas to which they pertain (references on
>last page).
>
>DESCRIPTION
>Flow cytometry is a  technique to analyze single cell suspensions from a
>variety of sources including blood, bone marrow, body fluids, lymph nodes,
>and other tissues. A variety of cell surface and intracellular antigens
>are detected with the use of fluorescent probes and fluorescently-labeled
>antibodies. Flow cytometers contain lasers, fluidics, and optical
>detection systems to rapidly and sensitively detect fluorescent signals
>from a large number of cells. Detection of specific protein antigens and
>quantitation of cellular DNA can be useful in the diagnosis, monitoring of
>response to therapy, analysis of relapse, detection of minimal residual
>disease, and as indicators to predict prognosis in a wide variety of
>disease states.
>HCPCS CODES (Alpha numeric CPT © AMA):
>Code: Descriptor:
>88180 Flow cytometry; each cell surface marker
>88182 Cell cycle or DNA analysis
>
>Comments:
>        CPT code number 86361, which is not on the above list, should be
>used for CD4 absolute numbers.  Both the percentage and absolute numbers
>of CD4+ CD3+ cells should be reported (1).
>
>INDICATIONS:
>1. HIV Infection. Determination of the numbers of lymphocytes in the
>peripheral blood is useful for the diagnosis and monitoring of patients
>with HIV infections. Testing of other body fluids and tissues may also be
>useful in selected clinical conditions. The most clinically relevant
>marker to be measured is the number of CD4+ lymphocytes, [But there have
>been questions raised that the standard of practice should include CD3,
>CD8, CD16 and/or CD56, and CD18 with CD 45 and CD 14 used as part of the
>analysis gating strategy. Please address comments to this.]. The
>lymphocyte analysis is initially done as part of the diagnostic workup and
>monitoring in stable individuals at no greater than 3-month frequency
>intervals. Prior to initiation of anti-viral therapy the test may be
>performed on two different occasions to ensure accuracy of the results and
>establishing good baselines. Once therapy has begun, follow-up tests may
>be done at 1 month, and then done every 3-4 months thereafter. Testing may
>also be appropriate when a significant change has taken place in the
>patient's clinical condition.
>
>Comments:
>        I agree that the most clinically relevant number to report is the
>CD4+ lymphocytes, but the population must be confirmed to be CD3+ and
>CD8-.  The use of CD45 as an additional marker is critical in the strategy
>of gating the appropriate population (1).  Our lab runs a single tube four
>color analysis to include 3/4/8/45.  Reimbursement should compensate for
>at least 4 antibodies.
>
>2. Leukemia or Lymphoma. For hematopoietic neoplasms (leukemias and
>lymphomas) immunophenotyping is an important adjunct in primary diagnosis,
>staging, assessing prognosis, and, in turn, formulating an optimal
>treatment plan. Furthermore, it occasionally serves as a tool in
>monitoring response to therapy.
>Primary Diagnosis and Staging: The characterization of cell surface
>markers is appropriate in suspected hematopoietic neoplasm, especially
>where first-line diagnostic methods fail to establish a diagnosis. In many
>cases, a panel of up to 12 antibodies (professional and technical
>components) will be adequate for the diagnosis of acute leukemia. In cases
>of lymphoma, a panel of 12 antibodies (for the professional and technical
>components) and up to an additional 6 antibodies for the technical
>component will be considered adequate for the diagnosis.  Documentation
>regarding medical necessity is to be provided for the use of additional
>antibodies.  However, simple diagnosis such as uncomplicated CLL may
>require only 6 markers for diagnosis.
>
>Comments:
>        I strongly disagree with a limit of 12 or 18 antibodies for the
>diagnosis of leukemias and lymphomas.  Any individual malignant population
>can probably be sufficiently defined by describing its level of expression
>of 12 separate antigens, as in the tables in the article cited in the
>recommended policy (2).  However, examination of the tables in this
>article shows that the antigens described are different for each disorder
>listed.  These neoplasms were certainly not detected using panels
>containing only the markers shown for each process!  In that case the
>nature of the neoplastic process would already have to have been known in
>order to have selected the appropriate 12 antibodies.  In practice, this
>is obviously not the case.  The indication for biopsy/marrow aspiration,
>etc., is often something like lymphadenopathy or cytopenia, and a large
>number of disorders are within the differential diagnosis.
>        The US-Canadian Consensus Conference has published recommendations
>regarding the immunophenotypic analysis of hematologic neoplasia by flow
>cytometry (3,4).  This  consensus conference included clinical
>hematopathologists, hematologists and laboratory scientists representing
>academic institutions, federal agencies, large hospitals and commercial
>diagnostic laboratories.  Corporate representatives, members of regulatory
>agencies and observers from European Countries were also present.  As a
>result of the conference, 5 consensus documents were produced.  I think it
>is appropriate to give more weight to the findings and opinions of this
>Consensus Conference than to a review article written by only two authors.
>        The consensus document that deals with selection of antibody
>combinations (4) provides lists of "core" and "supplemental" antibodies
>useful in immunophenotyping.  Although the lists of "core" antibodies can
>be seen to contain only 12 antibodies, it is stated in the text that "a
>screening panel serves only as an initial evaluation step that precedes
>the use of a more extensive set of antibodies necessary to complete the
>analysis".  The consensus paper also states that "in general, the larger
>the number of reagents, the higher the sensitivity of abnormal cell
>detection and the better the ability of delineating phenotypes that may be
>useful in disease monitoring" and "neoplastic cells often exhibit
>abnormalities in antigen expression, a feature that may be overlooked if
>an inadequate number of antibodies is used.  A restricted panel may also
>limit the ability of recognizing minimal neoplastic involvement."  In
>addition, the conclusions state that "The panels should be designed to
>resolve normal as well as malignant cells because normal cells act as
>internal reference standards.
>        In the consensus document, the Conference Committee quoted studies
>showing that the average number of reagents routinely used by North
>American laboratories engaged in the analysis of hematopoietic neoplasia
>is 19 and 16 for acute leukemia and lymphoma analysis, respectively (5).
>In Europe the number of antibodies recommended is even higher (6).  The
>conclusion of the Consensus Conference is that "An adequate
>characterization of the cellular components in a sample requires at least
>these many reagents or even larger panels, which unfortunately affect test
>cost.  However, such a characterization is often of significant clinical
>value."
>        Also, this method of reimbursement does not take into account the
>fact that antibodies cannot be used individually or in arbitrary
>combinations, and to set up an effective panel requires specific
>combinations.  One reason for this is because the "highest intensity
>fluorochromes for the lowest epitope densities should be used" (4).  Also,
>"a certain degree of redundancy in antibody specificity is always
>desirable" (4) because "neoplastic cells often lack antigens expected to
>be present on normal cells or are expressed in an aberrant manner" (4)
>The same antibody often needs to be used more than once, in order to
>combine it with different markers, and the laboratory absorbs the cost of
>this duplication, as well as absorbing the cost of the isotypic control
>antibodies used for each specimen.
>        As a reimbursement policy, I think it would be appropriate that
>any antibodies beyond a certain number (maybe 12 or 16) be reimbursed at a
>lower rate, but I think that, based on the above arguments, it is poor
>diagnostic practice to systematically limit every case to a very
>restricted panel of antibodies.
>
>They can also assist in staging of a patient (with the exception of
>Hodgkin's Disease) to confirm a primary diagnosis of a secondary site.
>Immunophenotyping is not usually part of evaluation of patients with
>granulocytosis, nor is it medically necessary for the diagnosis of chronic
>myelogenous leukemia (CML) which is not undergoing blastic transformation
>(i.e., blast counts of greater than 5% in the bone marrow differential and
>in the peripheral blood).
>Prognostic Markers: After a diagnosis has been established, a panel of
>cell surface markers can provide clinicians with useful prognostic
>information for an individual patient. For example, patients with CD20
>positive lymphoma are candidates for anti CD20 antibody (Ritixumab)
>treatment. Also CD10 (CALLA) and CD34 negative leukemia are known to have
>a poor prognosis.
>
>        The antibodies listed here are important in the diagnosis of the
>disorders mentioned and should therefore be run in the original panel.  It
>is important for the original panel to be as complete as possible because
>"This procedure saves time, and additional staining is only rarely
>required." (4)  Neoplastic cells are fragile and degenerate over time.
>Thus, it is best to do all the necessary stains on the fresh specimen
>instead of performing a very limited panel and then doing extra markers,
>which delays diagnosis and can put the specimen in jeopardy of degradation
>
>
>Monitoring Response to Therapy: Presently this is an infrequent
>application of flow cytometry and is at carrier discretion.
>
>3. Organ Transplants. Postoperative monitoring of covered organ
>transplants may be done by analysis of lymphocytes in the peripheral blood
>or tissue of patients. Analysis of cells may be useful to indicate early
>rejection, bone marrow toxicity during immunosuppressive therapies, and to
>help in the differentiation of infections from transplant rejection. A
>variety of cell surface markers and activation antigens can be used
>depending on the clinical condition and the organ transplanted. Periodic
>monitoring may be done at intervals determined by organ-specific
>protocols. Testing may also be done when the clinical condition is
>indicative of possible acute or increased severity of chronic rejection
>episodes.  Peripheral blood testing may also be done to monitor
>effectiveness of anti-rejection immunosuppressive therapy which may
>include OKT3, anti-thymocyte globulin, or other toxic agents. The number
>of circulating T-cells is usually determined by measuring the cell surface
>marker CD3. [Comments are invited to explain the medical necessity of
>other markers, e.g., NK cell surface markers, CD4, 8, 19, 16, 56]. A
>single test sample is usually sufficient; however, additional testing may
>be necessary for further evaluation when patients show partial or no
>response to the therapy.
>
>4. Carcinomas. DNA analysis of tumor for ploidy and percent-S-phase cells
>may be helpful for patients with early stage carcinomas (ovarian, breast,
>colon, endometrium, prostate, and carcinoma of unknown primary site).
>Prognostic information obtained from DNA analysis may be useful in
>directing treatment decisions in patients with localized disease (stages I
>and II). Additionally, DNA ploidy and DNA index may have prognostic value
>in predicting long-term, disease-free survival in advanced ovarian
>carcinoma. DNA analysis is usually performed one time after an initial
>diagnosis has been made.
>
>5. Primary & Secondary Immunodeficiencies. A variety of primary
>immunodeficiency diseases have been identified in which surface or
>cytoplasmic proteins are missing or have impaired function. Primary
>immunodeficiencies usually occur in infants and young children and can be
>a result of defects in T-cells, B-cells, granulocytes, or monocytes.  The
>selections of antibodies used should be based on the clinical
>presentation, for the flow cytometry method can be used to detect
>abnormalities of a variety of surface markers. The selection of antibodies
>used should be based on the clinical presentation of each affected
>individual and the abnormalities most likely to be present.  Some flow
>cytometry testing can be done to detect possible dysfunction present in
>white blood cells. These defects usually cause primary immunodeficiency
>and are detected in newborns and young children, but may occasionally
>occur in teenagers or young adults. The flow cytometry testing may include
>the analysis of PMN and monocyte generation of oxidative burst and
>phagocytosis, analysis of T and B cell mitogen responses, analysis of NK
>cell tumor cell lytic function, analysis of platelet activation antigens,
>and the analysis of intracellular calcium and pH cell changes in cells.
>The selection of testing should be based on the clinical presentation of
>each affected individual and the abnormalities most likely to be present.
>A variety of secondary immunodeficiencies have been identified in which
>the patient presents with recurrent infectious diseases. The
>immunodeficiency is related to the occurrence of another illness or
>condition. Several of the more common diseases which can cause
>immunodeficiencies are severe malnutrition, biotin, B12 or zinc
>deficiency, lymphoma, myasthenia gravis, myeloma, radiation or
>chemotherapy, chronic alcoholism, drug abuse, cancers, splenectomy,
>chronic viral illnesses, and chronic disease. If a secondary
>immunodeficiency is suspected, an analysis of quantity and types of
>lymphocytes present in the blood can be performed to assess immune status.
>This could include detection of B-cell, NK cell, and T-cell antigens
>including CD4 and CD8.  Flow cytometry testing for primary and secondary
>immunodeficiencies is usually done as an aid to the initial diagnosis of
>the patient but may also, in some conditions, be used to monitor the
>disease or response to  the therapy or following bone marrow
>transplantation to assess reconstitution.
>
>Comments:
>        I agree with the recommendations on the use of flow cytometry for
>immunodeficiency testing and I support consultation between the
>pathologist and the clinician to best choose which markers and or tests
>are required based on the clinical findings.
>
>6. Stem Cell Transplantation. Hematopoietic stem cells which are positive
>for the CD34 surface protein are capable of reconstituting bone marrow
>with all the important cellular elements. Measurement of the percent and
>number of CD34+ cells in the peripheral blood prior to stem cell
>collection can be done to determine the appropriate time for the pheresis.
>The CD34+ measurement can also be done to determine the actual yield of
>CD34+ cells in the pheresis product to ensure the presence of adequate
>numbers of stem cells in the product. The analysis should include the CD34
>antibody but may also include several additional antibodies used to insure
>the correctness of the analysis and minimize overestimation of the numbers
>of stem cells present.
>
>Comments:
>        I agree with the recommendations regarding stem cells especially
>regarding the inclusion of several other additional antibodies to insure
>the correctness of the analysis and minimize overestimation, given the
>critical importance of harvesting sufficient stem cells for these
>patients.
>
>7.Paroxysmal Nocturnal Hemoglobulinuria. Paroxysmal nocturnal
>hemoglobulinuria (PNH) is an acquired disease characterized by the
>development of an abnormal clone of precursor cells in the bone marrow.
>The cells produce abnormal white cells and red cells, which have abnormal
>function and are susceptible to lysis. Analysis of this abnormality by
>flow cytometry, in general, can be accomplished by analysis of two
>antigens on the red or white cells. Further documentation may be supplied
>to support medical necessity of up to four antigens. In general, this
>testing would be done for diagnostic purpose only.
>
>Comments:
>        The experience in our lab has been that flow cytometry is a very
>sensitive tool for the diagnosis of varying degrees of severity of PNH.
>Because of the different patterns of GPI-linked antigens on various normal
>cell populations (for example, only granulocytes and B-lymphocytes would
>be expected to express CD24 - negativity is normal in other cell
>populations), we include markers which will separate the various
>populations.  Our panel includes CD14, 59, 45, 33, 16, 55, 3, 24, 19.
>Although a more limited panel may possibly be effective, there is not yet,
>to my knowledge, sufficient data to support that conclusion.  That may
>change in the future, because this is a relatively new diagnostic modality
>with which we have had limited experience.
>
>8. Other. Flow cytometry can also be used to directly demonstrate
>anti-platelet, anti-granulocyte, lymphocyte, and red cell autoantibodies.
>These antibodies can be found in a variety of autoimmune conditions and
>cause anemia, leukopenia, or thrombocytopenia. Detection can be either
>direct (patient cells) or indirect (patient serum) and use anti-human IgG
>antibodies. The analysis may also include several additional antibodies
>used to insure the correctness of the analysis for the cells of interest.
>
>Conclusion:
>        For these reasons, I think the proposed reimbursement policy would
>lead to an unacceptably poor quality of flow cytometric diagnosis.
>Accurate diagnosis is important in avoiding over-treatment, which can lead
>to unnecessary morbidity and mortality and increased expense, or
>under-treatment, which can lead to disease recurrence.
>
>Sincerely,
>
>
>
>Deborah B. Aquino
>Assistant Professor of Pathology
>
>References:
>
>1.  Centers for Disease Control and Prevention.  1997 Revised guidelines
>for performing CD4+ T-cell determinations in persons infected with human
>immunodeficiency virus (HIV).  MMWR 1997;46:1-30.
>
>2.  Jennings CD, Foon KA.  Recent advances in flow cytometry:  Application
>to the diagnosis of hematologic malignancy.  Blood 1997;90:2863-91.
>
>3.  Braylan RC, et al.  U.S.-Canadian consensus recommendations on the
>immunophenotypic analysis of hematologic neoplasia by flow cytometry.
>Cytometry 1997;30:213.
>
>4.  Stewart CC, et al.  U.S.-Canadian consensus recommendations on the
>immunophenotypic analysis of hematologic neoplasia by flow cytometry:
>Selection of antibody combinations.  Cytometry 1997;30:231-5.
>
>5.  Hassett J, Parker J.  Laboratory practices in reporting flow cytometry
>phenotyping results for leukemia/lymphoma specimens:  Results of a survey.
>Cytometry 1995;22:264-81.
>
>6.  Rothe G, Schmitz G.  Consensus protocol for the flow cytometric
>immunophenotyping of hematopoietic malignancies:  Working Group on Flow
>Cytometry and Image Analysis.  Leukemia 1996;10:877-95.
>
>


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:35:26 EST