EGFP-F Followup

From: Maggie Kasten (kasten1@jeflin.tju.edu)
Date: Fri Oct 30 1998 - 11:04:21 EST


Dear Flow-ers,

This is a follow up post to one I made a month ago about background
obtained with the EGFP-F plasmid. Several people expressed interest in
this construct and so I hope this information is useful to them and
others.

Here's a review of our problem. EGFP-F is an EGFP construct targeted to
the cell membrane through the ras farnesylation sequence and, therefore,
should not leak out of ethanol fixed cells. However, flow cytometry
analysis of our ethanol fixed cells transfected with EGFP-F showed
significant increase in the fluorescence signal of the untransfected
population. The increase was so severe that we despaired of using this
construct in our flow cytometry experiments. 

Andy Beavis posted that he has seen similar results with EGFP-F. He was
also kind enough to recommend a different membrane targeted EGFP
construct, EGFP-Spectrin, that I received. In our hands EGFP-Spectrin
also showed increased background, although not nearly as severe as
EGFP-F. (Sorry Andy.)

I received an e-mail message from a Clontech research scientist about
the background problem. My thanks to the person who forwarded her my cry
for help. She said that Clontech had heard that some researchers
experience increased background when using EGFP-F with ethanol fixed
cells. She believes that the background is due to "short" ethanol
fixation and that the problem can be remedied by fixing cells for at
least 24 hr or by washing the fixed cells with PBS + 1% BSA. I'm a bit
skeptical about the short fixation explanation as we routinely fix our
cells overnight. However, washing the cells with PBS + 1% BSA did
eliminate our background while maintaining the brightly fluorescent
cells.  This addition worked to reduce the background for both the
EGFP-F and the EGFP-Spectrin transfected cells. It seems unlikely that
the BSA step will interfere with subsequent treatment of the cells, PI
staining for cell cycle analysis in our case, and so should be generally
useful. 

Thanks to all who responded.

Maggie Kasten
Flow Cytometry Newbie
-- 
Maggie Kasten			Voice: 215-503-4614
Pathology Department		FAX:   215-923-9626
Thomas Jefferson University	Philadelphia, PA


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:35:26 EST