Re: Naive Memory Lymphocyte enumeration

From: Mario Roederer (Roederer@Beadle.stanford.edu)
Date: Fri Oct 30 1998 - 11:44:41 EST


<c=US%a=_%p=Specialty_Labora%l=PRIMARY-981028233759Z-63810@primary.specialtylab
 s.com> of Wed, 28 Oct 1998 15:37:59 -0800
Content-Type: TEXT/plain; charset=US-ASCII

Susan, 

Identifying naive T cells in the human is nontrivial and you must take care to
do it properly!  In any case, you must use at least 3 colors to do so:  CD4 or
CD8, and a pair of antibodies including a CD45 isoform (as noted below).

There are several ways to uniquely identify T cell subsets in the human.
Currently, the best are the combined use of CD45RA with one of CD11a, CD62L, or
CD27.  Naive T cells are CD45RA+ and CD62L+, CD27+, and CD11a-dull.  

You may use CD45RO in place of CD45RA, noting that its expression is inverse to
CD45RA; thus, naive T cells are CD45RO- and CD62L+, CD27+, CD11a dull.

Don't make the mistake of using CD45RO and CD45RA in the same stain; you get no
additional information from using either one alone (see notes below for more
info).  Also note that CD45 isoform expression is very different on CD4 and CD8
T cells!  Therefore, you CANNOT use isotype gates for identifying naive T cells,
and you CANNOT use the same gate set for CD4+ cells as for CD8+ cells.  In
particular, the CD8+ memory cells that are "CD45RA-" actually express a good
amount of CD45RA!  You will need some experience with staining healthy adults to
see where to set the gates.

CD62L is probably the best antigen to use in combination with CD45RA.  It is
bright; the CD62L+ cells are easy to distinguish from CD62L- cells.  The main
problem with CD62L is that it falls off cells during freeze/thaw protocols, and
cannot be reliably used on frozen blood samples.

CD27 is a good choice as well, although far fewer functional studies have been
done with CD27+CD45RA+ cells to really prove that these are pure naive T cells.

CD11a is the most robust identifier of naive vs. memory cells in the CD45RA+
subset.  However, it is the most difficult to use, because you must set a gate
that distinguishes bright from dull cells, without that much separation.  Again,
a bit of experience looking at healthy adult samples will teach you quickly
where to set this gate.

CD45RA and/or CD45RO staining is insufficient without an additional marker
(CD62L, CD27, or CD11a).  There is a distinct set of CD45RA+CD45RO- memory T
cells.  In the CD8 compartment, this subset averages around 10-30% of total CD8;
in CD4 it averages 5% of total CD4 (in healthy adults).  In adults with acute or
chronic disease such as HIV, these percentages go way up.  In advanced HIV
disease, CD4 T cells can be more than 50% CD45RA+, but nearly all are memory T
cells!

In unpublished studies using 10-color flow cytometry, I have correlated the
exact expression of all of these markers (together with functional studies to
verify memory/naive as well as possible).  I do find that about 1% (CD4) or 5%
(CD8) of the "naive" T cells, when identified using CD45RA with CD62L OR with
CD27 are actually activated memory T cells!  If you use CD45RA, CD62L, and CD27
altogether (a 4-color combination with CD4 or CD8), then this percentage drops
by an order of magnitude.  Thus, a three color combination of CD4/8 with CD45RA
and CD62L or CD11a identifies naive T cells to about 95-99% purity.  CD11a was
much better, achieving nearly 100% purity.

mr