Hi Gary,

Here's a (slightly edited) quote from Phoenix Flow Systems' web site:

"FACScan Service Tip of the Month

[snip]
How many of you out there have had the main air pump replaced on your
FACScan? Did you know that the main air pump is rebuildable? Did you also
know that the parts to rebuild the main air pump come in every PM kit
ordered from BD? There are two parts in the PM kit with the following part
numbers: Valve set, P/N
06-30214-00 & Diaphram, viton, 06-30215-00. By replacing these parts in the
main air pump, your air pump will work like new."

I think I've got a couple of these spare, if you have trouble getting hold
of them, drop me a line and I'll send them to you.

Ray

>Hello Flowers,
>
>We have an older BD FACScan and it appears that the air pump is slowing down.
>Right now the instrument is operating fine but it may continue to get worse as
>time goes on.  We could simply buy a new air pump from the vendor but being
>without a service contract I always like to explore all the possibilities.
>Has
>anyone had any experience with replacing the stock air pump with an off the
>shelf model or connecting a FACScan directly to house air?
>
>Thank you in advance,
>
>Gary Ward
>Englert Cell Analysis Lab
>Norris Cotton Cancer Center
>Lebanon, NH


                              Ray Hicks
________________________________________________________________________
|University of Cambridge          |Tel              01223 330149        |
|Department of Medicine           |Fax             01223 336846         |
|Level 5, Addenbrookes Hospital   |e-mail         <rh208@cus.cam.ac.uk> |
|Hills Road Cambridge             |Web  http://facsmac.med.cam.ac.uk    |
|CB2                              |ftp server  ftp://131.111.80.78      |
|UK                               |                                     |
|_________________________________|_____________________________________|


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:35:24 EST