Re[2]: sterilizing sorters

From: Gerhard Nebe-von-Caron (Gerhard.Nebe-von-Caron@Unilever.com)
Date: Thu Sep 10 1998 - 16:08:26 EST


          Dear Colleagues
          
          Please consider the differences in instrument design of the 
          fluidics. If you for example fill up the sheath filter of an 
          Elite with bleach it will leak out of there for quite some 
          time.  Your best bet is to disinfect the pipelines somewhere 
          in the sheath line close to the flow cell and put in a fresh 
          sterile 'disposable' sheath filter right in the same place 
          after rinsing with etoh.  This is how I sort bacteria on 
          enrichment and recovery media w/o contamination with a 
          minimum of expensive operator handling (and otherwise 
          potentially wasted) time compared to the price of the 
          appropriate sterile filter. However, consider the potential 
          reduction of sheath pressure at the flow cell caused by such 
          a filter on your fluidics / sort-system behaviour.
          
          Regards
          Gerhard


______________________________ Reply Separator _________________________________
Subject: Re: sterilizing sorters
Author:  trotter@scripps.edu at INTERNET
Date:    08/09/1998 19:37


Tony,

   We use only ETOH (70% in dH2O), dH2O, or PBS in the sheath tank/line
system. However, I have used 10% bleach every so often in the sample
line for a minute or so, followed by 2 washes of dH2O, a wash of 70% ETOH,
and then backflushing with sterile (filtered) sheath buffer. We have never
had any sign of toxic effects, but I have only done the bleach wash to
remove what I believed to be biological debris in the line,
not to sterilize the system (and not very often). If something is
really wrong in the sample line we just replace it and use ETOH and
sterile PBS to wash it before putting on cells.

     The first thing we did once we saw an apparent toxic effect
to the cells once sorted was to 1) incubate them in different PBS
buffers for a few hours and 2) put them back in the incubator. The cells
grow just fine once exposed to the PBS we made and the BioSure isotonic
saline, but die (slowly) after exposure to the Criterion diluent 2. So,
we're talking about a toxic buffer, not a toxic sheath line. That lot
of Diluent 2 is now only being used in the FACScans.


               Regards,

               Joe

On Thu, 3 Sep 1998, Antony Bakke wrote:

> Joe,
>
> There has been some discussion on the purdue group lately about sterilizing
> including mention that ethanol does not kill spores, but bleach does.  Because
> of that I decided to switch to bleach, but unfortunately found that even hours
> of running sheath did not remove all the bleach, since the sorted cells were
 all
> dead.  How do you sterilize your instruments and if you use bleach, how do you
> insure that it is removed?
>
> Thank you,
> Tony
>