I got several replies from others who have tried to fix Annexin V stained
cells. One was positive and the others negative. I have cut and pasted the
replies below.

1. I have used annexin with fixation with reasonable results, see Manion
and Frey,
Cytometry (CCC) 26:317-322 fig 3 for data.  The combination of no wash and
fixation did degrade staining in that figure.  The key trick is to supplement
your fixative with divalent cations, we use FACSlyse with 1 mM calcium.  This
has worked in several studies for me.  A second key observation is that annexin
binding is reversible even after PFA fixation, any washes also need to be done
in calcium supplemented buffers.

2. Hello,We have tried fixing cells after annexin V staining, in PBLs. Our
experience is that the level of apoptosis that we see by annexin is less
after fixation than if we analyse cells that are freshly stained.
Therefore, if you need to use a fixation step, I suggest that you think of
using a tunel stazin, which in our hands is quite amenable to fixation.

3. We've not convinced ourselves that any fixation technique (particularly
EtOH) is compatible with Annexin-V staining. All pre-fixes have been a
bust, and post-fixing does not give us the same results. Our experience,
however, is limited to tissue culture cell lines- mostly ones growing
attached to plastic.

4. >I tried fixing stained cells following the protocol from Techniques in
Apoptosis: A User's Guide: T. Cotter and SJ Martin 1997, page 113, but
without good results.  As the authors say in that book, fixation procedures
will allow any free annexin v to enter the cell and bind to PS on the inner
leaflet of plasma membrane.  From my own experience, the best way to do it
is to have a small number of samples and to analyse them as fast as you
can.  Keeping them on ice alters the results as well.
>For detection by immunofluorescence microscopy in tissues, once you
>section the tissue, you will expose PS from the inner leaflet and get
>false positives.
>

Maryalice Stetler-Stevenson