RE: Sample Differential Failure

From: Steve Merlin (merlin@aitken.org)
Date: Tue Jul 28 1998 - 09:30:00 EST


Eric,

To eliminate moisture in the air line for the Vantage and protect the
sample differential regulator, I purchased a Drierite plastic drying column
from VWR Scientific, Tel : 1-800-932-5000 Cat. # 26668-007 (Mfg #26800).
The dessicant will last several days before needing recharging or
replacement. Refill dessicant is the 8 mesh size, Cat# 22891-040 (mfg #23005). The
cost is around $40.00 for the drying tube.

The other option is to use bottled nitrogen as the pressure source. I
found the drierite method to be sufficient to solve the problem and easy to
maintain.

Steven Merlin
Aitken Neuroscience Center
Flow Cytometry Facility
523 East 72nd Street/8th Floor
New York, N.Y. 10021
212-772-6426 (Lab)
212-772-0608 (Office)
212-772-0357 (Fax)
merlin@aitken.org

Eric Van Buren wrote:
>
>Last Friday we had the sample differential on our FACS Vantage fail. It
>appears to have "frozen up" due to some mineral deposits. Has anyone
tried
>to clean one of these? Is it practical?
>
>The source of the deposits appears to be our pressure pump. On humid
days,
>it blows moist air through the cytometer. Has anyone else dealt with
this
>sort of problem? I'm interested in hearing remedies before I "re-invent
the
>wheel".
>
>Thanks,
>Eric
>
>/\/\/\_ Eric Van Buren, aa9080@wayne.edu
>\ \ \   Karmanos Cancer Institute and Immunology & Microbiology
> \_^_/  Wayne State University, Detroit, Michigan, USA
>
>
>
>
>RFC822 header
>-----------------------------------
>
>Received: from flowcyt.cyto.purdue.edu (flowcyt.cyto.purdue.edu
[128.210.60.31])
>	by crane.prod.itd.earthlink.net (8.8.7/8.8.5) with ESMTP id UAA25187
>	for <merlin@aitken.org>; Mon, 27 Jul 1998 20:03:18 -0700 (PDT)
>Received: (from root@localhost)
>	by flowcyt.cyto.purdue.edu (8.9.1/8.8.7) id SAA05541
>	for cyto-sendout; Mon, 27 Jul 1998 18:50:13 -0500
>Received: (from bin@localhost)
>	by flowcyt.cyto.purdue.edu (8.9.1/8.8.7) id KAA03505
>	for kelley; Mon, 27 Jul 1998 10:55:08 -0500
>Received: from mail4.wayne.edu (mail4.wayne.edu [141.217.1.82])
>	by flowcyt.cyto.purdue.edu (8.9.1/8.8.7) with ESMTP id KAA03501
>	for <cytometry@flowcyt.cyto.purdue.edu>; Mon, 27 Jul 1998 10:55:07
-0500
>Received: from [146.9.72.237] ([146.9.72.237])
>	by mail4.wayne.edu (8.8.5/8.8.5) with SMTP id LAA10980
>	for <cytometry@flowcyt.cyto.purdue.edu>; Mon, 27 Jul 1998 11:54:03
-0400 (EDT)
>Message-Id: <v01540b01b1e25e7777f0@[146.9.72.237]>
>Mime-Version: 1.0
>Content-Type: text/plain; charset="us-ascii"
>Date: Mon, 27 Jul 1998 11:53:45 -0500
>To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
>From: aa9080@wayne.edu (Eric Van Buren)
>Subject: Sample Differential Failure
>X-UIDL: aa398ce790b747fefa974c2920c5ed16
>


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:35:21 EST