platelets:CD62:PFA

From: Peter Schroeder (peter.schroeder@medtronic.com)
Date: Tue May 12 1998 - 11:30:14 EST


Thanks for all the information on the CD62 and PFA fixing.  From some of
the responses I judged it would be helpful to post some of the info to the
group.  

The take home message was that fixing with 1% paraformaldehyde after the
staining with CD62 would give the same results as samples that were never
fixed but labeled and run immediately (K.A. Ault, Biocytex).  
  
Thanks Again,
Peter Schroeder
Center for Biomaterials Research
Medtronic, Inc.


Kenneth A. Ault M.D. made this interesting comment on "pre-staining":
 
You have to remember that although resting platelets have no P-Selectin
ontheir surface, they have lots of it in their alpha granules.  Thus, if
the
platelets become permeabilized so that the antibody can gain access to
theinterior of the platelet, you get lots of staining.  Heavy fixation
causes
some permeabilization, and I think this is what others have observed when
theyreport that fixation "causes activation".

Bioctytex sent me the following information regarding their platelet
activation kit sold by Alexis:

Paraformaldehyde fixation can be used in two different ways.


First way : Sample fixation before staining to store the sample. 
In this case cell membranes can be altered and an increase of MAb binding
can be observed.
We have studied the influence of 1% PFA fixation on PRP and we have
obtained the following results :

	PRP (n=30)	            1% PFA fixed PRP* (n=8)
CD61	51,000 +/- 9,000         	70,000 +/- 8,800
CD62	     < 300	                            100 to 700
CD42b	37,000 +/- 6,000	         42,000 +/- 10,000

*	PRP fixation with 1% PFA (vol/vol)
	Incubation 10 min.
	Two washes in PBS buffer (1,200 g - 10 min.).

As Cahill reported, platelet fixation induces an increased expression of
the antigen (I don't know if it is due to platelet activation or membrane
alterations).

Second way : Sample fixation after staining to differ cytometry analysis.
In this case paraformaldehyde stabilizes MAb binding with no effect on Gp
measure.
"PLATELET Gp" procedure includes this final fixation step. We have some
data to compare technical protocol including final fixation with technical
protocol without this fixation step and we did not observe significant
differences in quantitative values. Thus, don't be affraid to fix CD62
stained samples.