RE: Nancy Barney

From: Marco Antonio Fernández (mafernandez@mx2.redestb.es)
Date: Sun Apr 19 1998 - 16:36:37 EST


Hi, Nancy
I too have been doing PAIg over these years. My experience is that fixed
platelets before stain may be stored at 4º C up to 48 hours. False positive
or increased fluorescence could arise after this time. Once stained, I
recommend you to acquire samples as soon as posible, within 24 hours.

Good luck.

Marco A. Fernández
Biologist
Head Flow Cytometry Dept.
Reference Laboratory
Internet: http://personal.redestb.es/marcoafdez Clinical Cytometry
mafernandez@mx2.redestb.es
08024 Barcelona (Spain)


----------
> De: Nancy Barney <nbarney@nih.gov>
> A: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
> Asunto: 
> Fecha: sábado 18 de abril de 1998 17:06
> 
> 
> Hi Flow folks( anybody doing platelets) : A question from one fairly new
to
> Flow, we have been doing Platelet Associated Immunoglobulins for a few
> months, and the question arose how long can the fixed platelets stay
usable
> at 4 deg ? In other words once they are fixed ( not stained ) do they
have
> to be stained the next day ? Also once stained must they be acquired the
> same day or is there 24 or 48 hours leeway. Any info appreciated.
> 
> 
> Nancy Barney MT (ASCP)