Conjugation instability -Reply

From: T. Vincent Shankey (Thomas Shankey) (TSHANKE@wpo.it.luc.edu)
Date: Fri Apr 17 1998 - 19:42:18 EST


Keith,
Provided the fluorescein actually was conjugated to the antibody, the off
rate should be quite low. However, there is also the possibility that the
"conjugate" contains much free (unconjugated) fluorescein. You could
test this by running the "conjugate" through a small column of Sephadex
G-25 or G-50.
A final possibility (not exclusive of the above) is that your target
intracellular antigen is present in large amounts- consistent with your
flow results. A similar pattern is seen staining intermediate filaments, ie
cytokeratins. We have never been able to decrease the mean ch
fluorescence by mixing "unlabelled" antibody with FITC anti-cytokeratin
(this also assumes the labelled and unlabelled antibodies are equavalent).
In addition to the large number of target molecules, the antibody
manufacturese (and you know who you are) tend to go for high F:P
ratios - good for surface antigens not present in high copy number, but a
problem when you want to saturate (ie quantify) a high copy molecule like
cytokeratin.
Regards,
Vince Shankey