From: Howard Shapiro (hms@shapirolab.com)
Date: Wed Apr 15 1998 - 17:29:24 EST


David Dombkowski asks:
>
> I am interested in knowing whether I can power a Hammatsu R3896 PMT with
>E717-63 socket assembly with an Epics751 High Voltage power supply. Also, I
>would like to know if anyone is familiar with signal output from such a PMT
>and if I can convert output signal with MDADS ADC's and BD Data Lister
>ADC's. In general, any configuration information would be appreciated.
>Thanks.
>

As far as I know, the R3896 is a side-window tube with the same pin
configuration as an R928 or R1477, and you could presumably mount it
in the same socket as one of those tubes and use the same power supply.
What (I don't have my Hamamatsu catalog handy right now) is an E717-63
socket assembly?  If it has a built-in HV supply, you can't use it with
your Coulter HV supply, and if it has a built-in preamp, you can't use
it with your EPICS preamps.  Are you trying to add another tube to get
an additional color, or replace a less sensitive tube?  In general,
ADC's aren't going to be a problem; the output from all PMT's is
current, which is converted to voltage in the preamp well before 
signals get to the ADC.

If you need more info, call or walk over.

-Howard