>After being over with the 'attachment' subject, we're now starting on the
>'basic questions' one.  As a very new user, I allready apologize for the
>following question and hope that Howard won't find I'm too lazy (If it's
>the case, please direct me to the right source of information).
>
>So here's the basic one :  I'm using a FACSvantage and found that it sorted
>really bad on SSC and FSC.  I checked the alignement of the laser, I
>checked the position of the obscuration barre, I cleaned the objectives,
>and I optimized the SSC - Fluorescence dichroic.  What else could I do to
>optimize the signal on SSC and FSC ?
>
I don't think this is a "lazy"question, and, although others who post to
this list have far more hands-on sorting experience than I do, I'll put
my two cents in.  First, you'll have to tell me whether this is a problem
with bad sorting or with bad signals.  In general, a likely cause of
bad signals (and, potentially, bad sorting) on a droplet sorter like the
Vantage is too high a signal amplitude to the transducer; this perturbs
the stream enough to degrade its optical quality, accounting for bad SSC
and FSC signals, and would probably affect sorting as well.  The other
thing I can think of which would screw up scatter signals and sorting is
crud on the orifice, moving the core to an off-center position and pro-
ducing instability of the droplet breakoff point.  That's my song and
dance; let's see what the other experts have to say.

-Howard