re: low cell count specimen

From: Theresa L. Fecik (flowcytoqueen@juno.com)
Date: Tue Mar 03 1998 - 16:32:13 EST


Nancy,

If I have a specimen with a low cell count, and the pathologist REALLY
wants me to try to do flow cytometry analysis on it, then I will increase
the amount of specimen that I use per tube (if I have enough specimen to
do that).   We use less antibody per tube than the manufacturer calls
for, anyway, to save money--instead of 100 ul of blood/bone marrow and 20
ul of antibody per tube, we use 50 ul of blood/bone marrow and 10 ul of
antibody per tube (well.....O.K.--I use 100ul of sample and 10 ul of
antibody--but, it seems to work just fine).  For bone marrow and blood
specimens that have a WBC count of 3000 or less, 100 ul would be double
the amount of specimen for the amount of antibody being used, and I
wouldn't have to put the tubes through a second lysing cycle.   I
personally don't want to play with the amount of antibody any more than
it has already been played with.

Hope this helps.

Going with the flow,
Terry Fecik
Sinai Hospital of Baltimore