Re: absolute cell counting by flow.

From: Marcus Reckermann (recker@ftz-west.uni-kiel.de)
Date: Wed Feb 25 1998 - 07:57:08 EST


> Subject:       absolute cell counting by flow.
> Date:          Mon, 23 Feb 1998 23:38:58 -0000
> From:          Aki Hoji <akhst7+@pitt.edu>
> To:            Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>

> 
> Dear, 
> 
> I am looking for a way to accurately count absolute cell numbers (from a 
> limited number of cells) by flow.  I'd appreciate if anyone could send me 
> a working protocol to do this, or provide some refs.  Thanks in advance.
> 
Dear Aki,
well, depending on the instrument you use, you have different 
options. As far as I know, only instruments from Partec (and 
possibly BioRad) have hardware options for absolute volume counts. 
The mostly used BD and Coulter instruments donīt have that (as far as 
I know), and you have to resort to indirect methods: 

- you can add a known bead concentration, and then calculate 
the event number of your cluster relative to the known bead 
concentration in your sample. BD distributes so-called TruCount Beads 
that facilitate this procedure. 

- if you have a reliably stable sample flow (check that thoroughly!), 
you can relate the elapsed time with the measured volume.

- I use another trick: I fill the fluidics system with the sample to 
be measured, then stop the sample flow. Then I fill a tube with the 
exact volume I want to measure, and mount that onto the sample 
holder. I start the measurement simultaneousely on the instrument and 
on the aquisition software; I stop it simultaneousely when the entire 
volume has been pushed out and I can see air entering the exit 
tubing (which must reach down to the very bottom of the sample tube). 
This only works when you have a transparent sample and exit tubing. 
There is no dead volume in the sample line because I fill the system 
with the sample before I start measuring.

- another fairly simple method is to weight your sample tube before 
and after your measurement. You just have to calibrate a sample 
volume/weight ratio for each sample type.

I am curious to learn more ways of doing absolute volume counts!

Good luck,

marcus

__________________________________________________
Dr. Marcus Reckermann
Research- and Technology Centre Westcoast (FTZ)
Hafentoern
D-25761 Buesum
Germany
Tel: +49-4834-604-204 or -261
Fax: +49-4834-604-299
E-Mail: recker@ftz-west.uni-kiel.de


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:35:13 EST