Jodi

While Brian makes a useful suggestion for cells, the protocol that you are using
will probably prevent it from working.  The NP-40 you use to make nuclei will
probably also extract any membrane label.  If you feel that you MUST make nuclei
you will likely be limited to shooting (in the relative dark) for an antibody
that recognizes an internal antigen retained by the process.  A species specific
nuclear structural protein might work, or if you are lucky some type of
intermediate filament in the stromal cells might be preserved.  Usually this
type of prep leaves enough cytokeratin in epithelial cells, but my recollection
is that stromal cells are endothelial or mesothelial - perhaps a vimentin or
something.

A less random thing to do might be to use live cells, UV exictation, and Hoechst
33342.  You could try running the cells and see if they are distinguishable by
scatter when unfixed to determine if this is a useful method to pursue. (The
suggestion of membrane or MitoTrackers would also work in this system.) If you
don't have UV - or if you think that H342 doesn't give suitable CVs - you could
try ethanol fixation.  Some surface eptiopes are preserved by this process, you
might be able to use CD38 or some mouse stromal marker, or even cytoplamic Ig.

Just some thoughts.

Tom Frey


_______________________________________________________________________________
Subject: RE: Cell cycle analysis on myeloma cells
From:    "Newsom; Brian S." <bsnewsom@msmail.his.tch.tmc.edu> at INTERNET
Date:    2/20/98  2:47 PM



Jodi,
One thing you might try is to label the myeloma or fibroblast cells with   
a lipid dye such as PKH. These dyes remain intact in the cells for up to   
100 days depending on cycle rate. As the cell divides and becomes two   
daughter cells the dye intensity is cut in half due to division of the   
membrane between the two new cells. I would say choose the cell that   
cycles the slowest and stain it with this dye. It comes in green and red   
and maybe other colors and works well and does not effect cell viability.   
It should make it easy to distinguish your two cell types.

Brian Newsom
Flow Cytometry Specialist
Baylor College of Medicine
Shell Center for Gene Therapy


You Wrote:

Dear Colleagues,

I am submitting this question for a researcher who uses my core facility.
Objective:  To study the cell cycle in human myeloma cell line(s) plated   
on top of a murine stroma cell line.
Hypothesis:  Upon adherence of myeloma cells to stromal cells, the cell   
cycle distribution of myeloma cells may be affected.
Procedure:  Human myeloma cells are plated onto the monolayer of   
gamma-irradiated murine stromal fibroblasts.  After 2 hours, non-adherent   
myeloma cells are removed.  Adherent cells are further cultured for up to   
72 hours.  After completion of co-culture, cells are lysed/stained in   
situ with a buffer containing 0.1% Na citrate, 0.3% NP-40, 100ug/ml RNase   
A, and 50ug/ml propidium iodide, vortexed at a high speed for 1 min., and   
kept on ice until flow analysis.
Problems:  We cannot consistantly discriminate myelomas from stromas   
based on their morphologies.  In each of four experiments, the stromas   
have varied in size and complexity (why I don't know, and the researcher   
cannot explain this issue).  However, the myelomas have consistantly   
shown the same size and complexity.  Therefore, we cannot use morphology   
to distinquish myeloma from stroma because I cannot single out the   
myeloma with a gate.
The researcher attemped to pre-stain the stroma cells with a SYTO dye,   
but the fluorescence of this dye was completely quenched upon PI staining   
as above.  (Has anyone seen this happen or have an explanation for it?)   
 The initial reasoning for doing this was to see if we could gate out the   
green fluorescent stromal cells from the PI stained myeloma cells.  (I   
wasn't sure if this was going to work, but I thought it was worth a try.)
Questions:  Has anyone successfully done anything like this in the past   
who might be so gracious as to share his/her protocol?  If not, does   
anyone have any suggestions as to what my fraid researcher might try?   
 Sorting the two populations does not seem like a "viable" option because   
the myeloma are already in a fragile state after culture.
Any response is greatly appreciated at this point.