intrinsic versus extrinsic -Reply

From: Antony Bakke (bakkea@ohsu.edu)
Date: Thu Jan 22 1998 - 15:18:01 EST


Andrea at Dahl-Chase asked:

Is washing with PBA sufficient to ensure that immunoglobulin staining is
intrinsic and not extrinsic (cytophilic)?
------------------------------------------

I do not believe that a single wash with PBA (I assume this is PBS with BSA and
azide) would be sufficient.  Cytophilic Ig is removed either by an acidic wash
or by incubating at 37 C for 20-30 min. and then washing with warm media 1-3
times.  We incubate ficol/hypaque purified mononuclear cells at 37 C in RPMI
with 10% FCS for 30 min and then centrifuge through a cushion of fetal calf
serum.  On occassion I have just washed F/H cells or lysed whole blood once with
PBA and there was still cytophilic Ig present.

Tony Bakke, PhD
Director, Clinical Immunology and Flow Cytometry Lab
Dept of Pathology
Oregon Health Sciences University
Portland, OR
bakkea@ohsu.edu