Re: CD34+ subsets & CFU-GM

From: Scott Burger (burge009@gold.tc.umn.edu)
Date: Fri Jan 16 1998 - 14:49:59 EST


Dear Dr. Vogt:

A few years ago, our transplanters asked for CFU assay results on almost all
products.  We have supported this, doing a progenitor assay on any cell
product we process.  The logic here makes some sense -- flow cytometry, cell
counts, and viability measurement are all invaluable, but do not evaluate
the functional characteristics of the cells.  The progenitor assay allows
one to assess the proliferation and differentiation functions of the cells,
which is what you care about in transplant in the first place.  So, I'm
pretty strongly biased in favor of CFU measurement.  

As CD34 measurement has improved, however, our transplant docs now are
saying that they are less interested in progenitor assay results.  I'm the
one who isn't comfortable with this, but it is an increasingly common
viewpoint among transplant clinicians.  While we still do the assay and also
get flow cytometry results, I know that a number of transplant centers now
do CFU's only under special circumstances.  

I'm not surprised your reference lab doesn't offer CFU assays.  It's a
fairly time-consuming test, and I'm not sure it would be one of the more
profitable.  I'd be interested in knowing the preferences of your transplant
docs on this one.  

Scott


***************************************************************