To answer a few questions from David L. Haviland:

I analyze calcium flux in lymphocytes on a regular basis on a cytometer,
and while tough, it is doable.  I do not think that any of the physical
differences you mention are relevant; in fact, I have compared dose
response curves of a  tumor infiltrating lymphocyte (TIL) line to a
chemokine, and they were identical.  

First of all, fluorimeters use ratio measurements, either fura (dual
excitation, single emission) or indo-1 (dual emission, single
excitation).  I cannot overstate the importance of ratio measurements.
The spread of loading within lymphocytes is quite large, but on a bound
vs unbound probe scattergram, the events are a very tight line.  If one
were to look at either parameter of the ratio alone, there is no way you
could detect a small number of responding cells, or even a small
response of a large number of cells.

The calcium rise after ionomycin is huge, and much greater than any
physiologic ligands I have tested (cross-link BCR, TCR, or chemokines).
It is a good place to start to let you know that cells are loaded, but
not worth anymore than that.  With Jurkats, it is better to look at
cross-linking the TCR.  

Cells do move in a fluorimeter, there is a stirrer; a fluorimeter is
looking at many events at once, and there may be some amplification
there.  At least I think so, but those with greater knowledge of the
physics may know better.  It is possible that a small number of events
can be seen in a fluorimeter, if the signal from those events is high
enough.  One example is that many studies show "PBMC" responding to   a
particular chemokine.  In fact, it is only the monocytes, which make up
about 25% of the total.  

Fluorimeters are cheaper and easier to use than a cytometer by far.  If
one has a homogeneous population of cells (e.g. transfected 293) , a
fluorimeter is the best choice, because there is no reason one would be
doing subset analysis. 

Finally, molecular probes has a probe, I think it is "Fura red" that
increases in fluorescence when not bound to Ca.  It emits at 670nm.  You
can load your cells with fluo and fura red, and use a program such as
multitime to ratio the two signals if you want.


Ronald L. Rabin, M.D.
Laboratory of Clinical Investigation
National Institute of Allergy and Infectious Diseases
National Institutes of Health
Building 10/Rm 11N228 MSC 1888
10 Center Drive
Bethesda, MD   20892-1888
Phone (301) 402-4910
FAX     (301) 496-7383
email:  rr84g@nih.gov
 


This archive was generated by hypermail 2.1.6 : Thu Jan 01 2004 - 17:34:06 EST