>   I would like to get a feeling for the effect of BD's Turbosort option on
>the viability of different cell types:  human, mouse, rat; PBL's, LN,
>thymocytes,splenocytes, bone marrow etc.;  mature/immature, healthy or
>compromised etc;   hybridomas, transfectants etc.  Any and all info would be
>welcome. Thanks,
>
>Claude

I can't say about the Turbosort since we chose to buy Cytomation MoFlo.
However, we have been running regularly at 65psi (that's with a 70um tip and
100K+ drops/sec and 12-20Kcells/sec and 2 lasers).  Everything we have put
through has had excellent viability. The cells range from thymic and splenic
(mouse) lymphocytes,cytokine expanded  human dendritic cells,  to a wide
array of in vitro cell lines.   Some of the experiments have used the sorted
cells for RT-PCR with excellent results.  Many of the experiments involved
single cell cloning with excellent cloning efficiency.  Part of the success
may be due to Cytomation's specially desighned tips.

Larry Arnold