We have recently gotten into the Rhodamine 123 business, using it in 
conjunction with PE, ECD (PE-Red610) and PE-Cy5.  We find that it 
requires a good deal more compensation against PE than does FITC, but 
is perfectly workable.

Another thing worth mentioning:  Rh123 has been used to measure 2 
different things, which seem sometimes to be confused (or at least 
confusing) in the literature.  When used to stain cells under 
conditions that permit dye efflux (i.e. stain, wash, incubate @ 37o), 
it can be used to measure activity of the MDR gene product 
p-glycoprotein which actively pumps the dye from the cell.  When used 
at higher concentration, or under conditions where efflux is blocked, 
Rh123 has been used to measure mitochondrial activity (since this 
charged low molecular weight lipophilic dye preferentially binds to  
mitochondria).


    