We also see this dropping of the side scatter signal on our Facscan. It happens rarely but it is annoying. I agree with Nancy - if it were a physical problem like a clog or bubble, then all signals should be affected. To me, it looks more like communication- the side scatter detector "shuts off" for a second. The other reason I think this, is that it is exactly the same type of pattern every time- the signal drops straight down. I don't think a clog or bubble would be that reproducible. We also recollect the sample rather than gate. Hope we find the answer! Happy New Year MaryAnn DeMaria NERPRC Harvard Medical School