Re: FACScan side scatter

From: Vera Svobodova (SVOBVE@novell1.dept-med.pitt.edu)
Date: Wed Dec 27 1995 - 23:21:31 EST


RE:  FACScan side scatter "blips"

We also see the "blips" in SSC.  Our DB service engineer and us came 
to a conclusion that they are caused by microbubbles (we agree with 
Judy).  They are NOT caused by electrical surges (we have a good 
voltage regulator in line).  We do not deal with them during acquisition 
(keep acquiring the sample) but we "cut"/gate them out in our analysis 
(make a plot of SSC vs TIME (Winlist)). That way you do not have to abort long 
acquisition and lose the sample.     

Vera