Re: Old Parformaldehyde

From: /G=Gerhard/S=Nebe-von-Caron/OU=1890CHPE/O=TMGB.URC/@LANGATE.gb.sprint.com
Date: Wed Nov 08 1995 - 06:30:00 EST


          My memory worked and I found the note I made on the last 
          ISAC.  We had some discussion about fixatives in the cell 
          proliferation workshop. Old PFA builds up formic acid when 
          degrading. Apart form people who used commercial reagents 
          from flow manufacturers, somebody mentioned that 
          Polysciences sells "fresh PFA" alliquoted probably under 
          argon or so. This would be similar to the approach described 
          by Calman, so probably more expensive.
          
          Two other things I picked up at the meeting were to check 
          osmotic strenght of the PFA solutions and the fact that 
          formalin usually contains 15% MeOH for stabilisation.
          
          Gerhard Nebe-v.Caron
          Unilever Research, Colworth Laboratory Sharnbrook, 
          Bedfordshire
          GB - MK44 1LQ
          Tel:    +44(0)1234-222066
          FAX:    +44(0)1234-222344
          E.mail: gerhard.nebe-von-caron@urcgb.sprint.com
          
          
          
          
           


______________________________ Reply Separator _________________________________
Subject: Old Parformaldehyde
Author:  Calman_Prussin@d10.niaid.pc.niaid.nih.gov at INTERNET
Date:    08/11/95 03:49


I think it depends on what you are using the paraformaldehyde (PFA) for. The 
constraints on using PFA for fixation of already stained cells (for later 
analysis) is less than that for uses with intracellular staining. For the 
former, I think PFA in the frig for months should be fine.
          
A lot of the fears may simply be that, however a number of investigators I have
          
talked to make up fresh PFA for each use. I think the concern is old PFA causes
          
overfixation leading to increased nonspecific binding.
          
Considering how much trouble it is to make fresh PFA, we make up a couple of 
litres at a time, aliquot them into 15 ml tubes and freeze the whole lot. I 
will use thawed PFA for a week or two after thawing. I have heard that Ulf 
Andersson's group in Stockholm use a similar approach, except they don't hold 
on to the thawed PFA for 2 weeks.
          
Calman Prussin
Laboratory of Allergic Diseases
NIAID, NIH
          
---------------------- Replied Message Body ---------------------- 
Date: 11-2-95  6:48am
From: {herro001@maroon.tc.umn.edu}:unix:niaid
  To: calman prussin:10:niaid,
    karin hartmann:10:niaid,
    barbara butcher:4:niaid,
    ramya gopinath:4:niaid
Subj: Old Parformaldehyde
Also-to: cytometry mailing list <cytometry@flowcyt.cyto.purdue.edu>
          
--------------------------------------------------------------------
          
I have read the warnings to only use "fresh" parformaldehyde for fixation, but 
I
don't recall ever hearing why.
          
Could someone explain the sorts of artifacts that old paraformaldehyde will 
introduce?  And what exactly IS old?