background subtraction

From: Alice L. Givan (Alice.L.Givan@Dartmouth.EDU)
Date: Tue Nov 07 1995 - 10:13:05 EST


Here's an old problem:  What do people do about subtraction of isotype control
values from flow determination of the so-called "% positive" population?  In
most lab with which I am acquainted,  if the isotype control marker is set at
3%,  then 3% is subtracted from every subsequent sample.  Although I have never
heard it  discussed in public,  I have known for a long time that this approach
is  obviously incorrect -- as can be seen if one imagines a sample with clearly
ALL the cells brightly stained.  By subtracting the control,  it would then
appear that only 97% of the cells were stained.  The problem with this
"classical" approach is that we should be subtracting 3% of the true negative
cells only because those are the ones that are mistakenly contributing to the
determination of the positive population. 

When I follow this through algebraically,  I come up with the following
formula:

y=(x-c)/(1-c/100)

where y is the actual % positive
x is the measured % positive
c is the % positive set for the isotype control

Am I wrong about this (or is it so obvious that everyone else does it already)? 
Does someone have the patience to check my derivation?
Thanks.