Caltag fix/perm "kits" work fine. Some loss of SSC is generated by the 
perm reagent. We have used the MPO/Lactoferrin, MPO/CD3, and MPO/CD22 
cocktails by Caltag. These cocktails also work with BD's FACSLyse 
reagent. BD's anti-CD3 conjugates stain cytoplasmic antigens well.
Kirkegaard-Perry's anti-kappa or -lambda stain cytoplasmic antigens well, 
while the surface has been tagged with CD38(PE) and CD45(CY5-PE). We have 
not observed cytoplasmic CD22 after the surface has been found to be 
negative. I don't know of other mfgs of anti-MPO. NH4Cl treated specimens 
tend to have more loss of SSC than whole blood. Even if the SSC is pushed 
up to match lymphocytes (or blasts) for region analysis, separation (with 
regard to light scatter)
between myeloids and monos and blasts and plasma cells and whatever else 
is there is less. 

======================