Re: Weird DAPI observation

From: /G=Gerhard/S=Nebe-von-Caron/OU=1890CHPE/O=TMGB.URC/@LANGATE.gb.sprint.com
Date: Fri Oct 06 1995 - 11:38:00 EST


          As I know you have an Epics Elite, I can confirm having seen 
          similar problems with lipophilic stains like bis oxonol that 
          also labels up the tubing. It is a question of dye 
          concentration, push time and the cleaning process of the 
          tubing. It is worth therfore checking the backflushing time 
          and looking at the time versus fluorescence display.
          
          Gerhard Nebe-v.Caron


______________________________ Reply Separator _________________________________
Subject: Weird DAPI observation
Author:  hilliard@cb.uga.edu at INTERNET
Date:    05/10/95 21:40


Hi folks,
          
This one has me scratching my head--we're using DAPI to stain for DNA 
content, and our first sample always has a lower fluorescence than the 
subsequent samples.  We thought at first it was staining variability, 
but out of curiosity we re-ran the first tube at the end of the 
experiment, and it had gotten brighter, so that all cells had similar 
DNA content (as we were expecting).  This has been consistent through 
the last 3 visits.
          
Is this some sort of tubing/sheath equilibrium effect?  I hadn't 
heard of this with DAPI, but we haven't done that much with it.  has 
anyone seen this before?
Steve
          
********************************************************* 