Re: Dual staining with Hoescht 33342 and PI

From: /G=Gerhard/S=Nebe-von-Caron/OU=1890CHPE/O=TMGB.URC/@LANGATE.gb.sprint.com
Date: Fri Oct 06 1995 - 04:52:00 EST


          pi will quench the hoechst fluorescence when they come in 
          close proximity by way of energy transfere (see Practical 
          Flow Cytometry by Howard Shapiro second edition page 169). 


______________________________ Reply Separator _________________________________
Subject: Dual staining with Hoescht 33342 and PI
Author:  MORGANTL@aol.com at INTERNET
Date:    04/10/95 21:58


Has anyone had experience with the dual staining method described by Belloc 
et al (Cytometry 17: 59-65; 1994)?  We are using a B-D FACStar Plus with an 
argon laser tuned to 310 nm.  We are using the filters suggested by Belloc:
 485/22 nm band pass and 600 nm long pass.  As postive test cells we are are
using a Burkitt's lymphoma cell line (Ramos) treated with H7.
          
Our problem comes when we add both dyes together.  We are able to detect blue 
fluoroesence from the Hoescht 33342 alone. However, when we add PI, the 
Hoescht channel PMT voltage has to be readjusted above 700 volts to achieve 
even a minimal ratio signal.
          
Thanks in advance for any help and advice.
          