Hi fellow sorters,

I have followed the exchange of data on cell rate versus detection rate for
a couple of days now and I feel compelled to suggest that the 'mailers'
read the article 'Coincidence in high-speed flow cytometry: models and
measurements' in Cytometry 12:398-404, 1991.
Yes, I realize it's my own work and it was done with a non commercial
instrument, but I believe it's relevant here.

The article describes several models that allow the analysis and prediction
of coincidence in the instrument 'dead time' (or 'cycle time' or
'interrogation' time as BD calls it) as a function of the dead time and the
cell (true throughput rate).

Lately, I have been trying to COMPLETELY understand what is going on in a
FACStar Plus as I want to get maximum performance out of it for CD34 sorts.

I have done similar measurements with yeast cells (they don't stick to
tubing) and I have obtained similar results. It should be noted that all
numbers represent the output of counters.


                     ANALYSIS RATES


                  CELL RATE vs SYS RATE
------------------------------------------------------
 cell rate (/s)        sys rate (/s)    efficiency (%)
    245                     245             100
    490                     508             104
    980                    1028             105
   1960                    1970             100
   3920                    3970             101
   7840                    7223              92
  15680                   11636              74
  31360                   16131              51

N.B. The interrogation time was set 20 us.

As the FACStar aborts a the processing of a trigger pulse when a 2nd event
is detected within the interrogation time, the fraction of detected single
cells (=sys rate) can be approximated from:

f(1)= P(1) / (1 - P(0) or roughly f(1)= 0.5 * PW *CR,

where PW is the pulse width (20 ms) and CR is the true cell rate (cells/s).


                        SORT RATES


                    CELL RATE vs SORT RATE
------------------------------------------------------------------
 cell rate (/s)      sort rate (/hr)   interrogation (us)  eff.(%)
    490                     20,220         20              100
    490                     19,620         40               97
    490                     18,240         60               90
    490                     21,900         80              108
   3920                    119,400         20               74
   3920                    122,400         40               75
   3920                    150,180         60               65
   3920                     90,540         80               56
  15680                    188,520         20               29
  15680                    187,620         40               29
  15680                    114,300         60               17
  15680                     70,440         80               11

N.B. The sort was NORMAL-R, drop drive was 37 kHz (PW = 81 us), 3
drops/cells, and the sort fraction was 0.017.

At the high cell rates there is a dramatic drop in performance that cannot
be easily explained. I am still in discussion with Larry Duckett from BD
about what all the aborts are in the FACStar Plus and how they are
processed. It tricky.


                    SORT MODE vs SORT RATE
-------------------------------------------------------------

 cells   ENRICH (3)  NORMAL-R (3)  NORMAL-C (3)  NORMAL-C (1)
 (/s)      (/hr)       (/hr)         (/hr)         (/hr)
    245     9,720       8,880         9,540         9,120
    490    19,620      18,540        18,660        18,360
    980    39,240      40,380        30,240        34,020
   1960    70,620      65,280        42,780        60,720
   3920   139,020     123,660        51,180       101,820
   7840   227,940     185,100        32,460       127,380
  15680   256,860     186,240        10,140       105,120
  31360   160,680     111,900           660        42,000

N.B. Interrogation time was 20 us, the sort fraction was 0.017, the drop
frequency was 37 kHz and (1) and (3) represent 1 drop/cell and 3
drops/cell, respectively.

>From these data, rough estimates be made of the sort purities when the
NORMAL-R and ENRICH mode are used.


My conclusion was to give up trying to make a slow sorter sort at
high-speed by adopt a combination of immuno-magnetic bead enrichment and a
slow (< 1000/s) final purification sort with coincidence checking on and
pulse processing activated.
For applications that do not allow pre-enrichment by surface markers, the
FLU trigger method is the way to go as it saves a lot of time that is
normally wasted on baby-sitting a long sort.

Hope this is some help.

Jan