In regards to Josh Cherry's question about autocloning into 96 well plates on an EPICS 752- We have routinely used the autocloner on our EPICS 753 for single and multiple cell deposition into 96 well plates with good results. The autocloner manual says that it takes the instrument 3.2 minutes per plate to transit all 96 wells. The time needed for the instrument to deposit a single cell when the sort population is at least 2% of the total does not significantly add to the total sort time for each plate. We usually allow around 5 minutes per plate when scheduling cloner time. Hope that helps. Justin Fishbaugh University of Iowa Flow Cytometry Facility internet: system@flocyt.int-med.uiowa.edu