reply to Salmon

From: Christopher A. Worth (CAWORT01@ULKYVM.LOUISVILLE.EDU)
Date: Wed Aug 18 1993 - 14:55:44 EST


Biomedical Engineer, BCC
Phone: (502)588-7191
I think your problem is that they dye is leaving your cells to get to an equili
brium point with the system.  When I was using PI-DNA analysis on a regular bas
is, i found that the G0/G1 peak would drift right as the samplestarted running.
What you may be experiencing is this same thing, however it is taking longer si
nc you are using a different dye.  The 7AD might just bind more tightly to the
DNA and it takes longer for the eq point to be reached.

My two cents,

chris