We're just starting to do large particle analysis/sorting with a  
FACStarPlus equipped with MacroSort and the 400um diameter orifice.  
After fitting the 400um nozzle, lowering sheath pressure to about 2  
psi, and realigning the instrument using chicken RBCs, we get the  
expected scatter (FSC and SSC) and fluorescence distributions  
(albeit, less precise). At a pressure of 2 psi, the measured stream  
velocity is about 2 m/s, which correlates well with the expected FSC  
pulse width. 


Corn pollen (from PolySciences), about 90um in diameter, was used as  
a large test particle. The roughly 60us pulse width for pollen  was  
as expected (beam height is assumed to be 20um at focus). However,  
there were substantial problems with clogging, hence, a sample tubing  
with larger internal diameter (and about the same outside diameter)  
is needed. Any reliable sources?

Further, sample settling is, of course, a problem and the informal  
documentation recommends against using sample agitate while sorting,  
but suggests the addition xanthan gum at 0.1% concentration. Any  
experience with this vis a vis, fluidics effects (viscosity of the  
sample stream will go up), sorting efficacy, sterility of sorted  
cells?

Any other gems of wisdom in sorting large particles on the FACStar?