Polarization on the FACStarPlus
Fluorescence polarization on the FACStarPlus is fairly easy to setup  
and measure.

There are two issues involved:
1. optics (getting, installing, and aligning)
2. data analysis: computing polarization from detector pulse height  
measurements.

Optics:
We've done fluorescence polarization using 488nm excitation. Optics  
include a 520LP filter (to remove 488nm side scatter), a 50/50 beam  
splitter (placed in the dichroic filter position between detectors  
FL1 and FL2) to separate vertical and parallel components, and  
polarizers in front of FL2 (vertically-oriented with respect to the  
plane of the laser) and FL1 (horizontally oriented).

The optical system was balanced using a half wave retarder in a  
rotation mount to adjust the vertical plane of the laser to 45  
degrees, and adjusting PMT high voltages such that the response from  
both detectors was equal. (The half wave plate was mounted on an  
aluminum plate fixed to one of the laser cover mounting positions in  
front of the 2W argon laser. Yes, the laser safety covers are  
removed. The cover mount on the cast optical bench is non-precision  
(that is, it is not parallel to the plane of the bench) so shims were  
used to get the half wave plate vertical plane perpendicular.)

Optics were purchased from Newport Optical (Fountain Valley,  
California) except for the 50/50 beam splitter mounted in an aluminum  
ring (special order; the beam splitter is very thin and fragile, you  
want it mounted in a ring) from Omega Optical (Brattleboro, Vermont).  
The beam splitter in its mounting ring was mounted in a  
custom-machined filter holder (working drawings available as a  
device-independent PostScript file by anonymous ftp at  
flowcyt.cyto.purdue.edu; see following article for instructions on  
downloading.)

Computation of Polarization

Fluorescence polarization is the difference of the parallel and  
perpendicular components divided by the sum of the two. Computing  
this value for each cell needs to be done using the parallel and  
perpendicular list mode values since there are no analog processing  
electronics on the FACStarPlus.

There are several ways to compute the polarization values from list  
mode data. We accomplished this using a simple package in Mathematica  
which computes the polarization for each value (multiplied by a  
suitable constant for proper scaling), and writes a new list mode  
file with the additional derived parameter.

A brief summary was presented in a poster at the last ISAC meeting.