Re: cox-2 antibody

From: Keith Bahjat (me@keithbahjat.com)
Date: Mon Nov 04 2002 - 14:06:39 EST


Jess,

Cayman Chemical has a FITC conjugated Ab that works. You'll need to do good
blocking to get it to work, though.

I recommend 2 blocking steps:

1) Dilute the Ab in 5% nonfat dry milk and incubate 30 min at 37 degrees
before using.
2) Resuspend the permeabilized cells for 30 minutes in 5% nonfat dry milk
before staining.

I've used BDs perm solution with good success in this protocol. Fix first
with mathanol free formaldehyde (Polysciences), then permeabilize with BD
perm, then resuspend in 100 uL of the 5% milk solution and incubate 30 min
at rt. Then add your Ab straight into this suspension. This worked well on
LPS stimulated human monocytes. Without the milk blocking steps, the
background is outrageous. This should work well with saponin as well, just
add a step with 5% milk in saponin staining buffer before adding the Ab.

Good luck.

----------------------

Keith Bahjat, Ph.D.
Scientist, Cancer Vaccines
Cerus Corporation
Concord, California