Re: Fluorochromes choice

From: Marty Bigos (mbigos@gladstone.ucsf.edu)
Date: Thu Oct 31 2002 - 11:34:03 EST


>Dear experts in flow cytometry,
>
>I have a possibility to use 6 colors FACS. The trouble is that I am
>not sure it is possible with the filters I have.
>I would very much appreciate any help.
>The instrument has three lasers: 325, 488 and 633 nm and I can use
>following filters:
>380/LP
>400/40
>424/44
>500/11
>510/20
>530/28
>575/15
>610/20
>660/13
>670/LP
>682/13
>Is it possible to find 6 different flourochromes, which would be
>efficiently differentiated with these filters/lasers ?

There are no useful fluorochromes for phenotyping that are excited by
325 nm. So you are limited to the 488 and 633 excitation.

In addition to bandpass detection filters, it is also important to
optimize the beamsplitters and the detectors, especially for
measuring the red emitting dyes.

Six dyes that would be useful for your system are:
Using 488 nm excitation:
FITC
PE
PE-TR (aka ECD from Beckman-Coulter)
PE-Cy7

and using the 633 nm excitation:
APC
APC-Cy7

We use these regularly on our Vantage SE for 6 color work.

However, without knowing your instrument configuration I could not
say whether they would work for you. I would guess that from the
filters you described above, you would have trouble detecting PE-Cy7,
APC-Cy7, and quite possibly APC.

Marty

>
>Thank you very much for your help.
>
>Michal
>
>--
>Michal Abel, MD, PhD
>Laboratory of Clinical Immunology
>INSERM U25
>Necker Hospital, 161 Rue de Sèvres. Paris
>tel: 33 1 42 19 28 87
>fax: 33 1 44 49 53 74


--
Marty Bigos
Director, Flow Cytometry Core Laboratory
Gladstone Institute of Virology and Immunology

Mail:
Box 419100
San Francisco, CA 94141

Packages:
365 Vermont Street
San Francisco, CA 94103

Direct Delivery (FedEX, UPS):
Gladstone Institute
San Francisco General Hospital
Building 3 Rm 509
1001 Potrero Avenue
San Francisco, CA 94110

(tel) 415-695-3832
(fax) 415-826-8449

mbigos@gladstone.ucsf.edu
http://gladstone.ucsf.edu



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:28 EST