Fanning - was GFP sorting question

From: Alan Bishop (alan.bishop@imvs.sa.gov.au)
Date: Thu Oct 17 2002 - 19:40:28 EST


Andy

The fanning you describe sounds very much like it is a problem caused by
small clumps of cells too small or loose to clog the nozzle orifice. As
these clumps pass through the nozzle orifice they are disrupted into single
cells (or smaller clumps), this disrupts the stable drop formation resulting
in the production of many drops of differing sizes.  The charge these a drop
carries is dependent on its diameter and carrying a different charge they
are accordingly deflected a different amount.

In the case of a specimen containing a high percentage of these clumps it is
possible to get what appears to be a continuous 'stable' fan, this fan will
occur even when no sort gates are selected provided the deflection plates
are turned on. You can check for this problem in two different ways.

1. Place a slide where you would normally place a tube, in bad cases you
will get a straight line with out any obvious indication of a spot, under
less extreme circumstances you will get the usual spot and a few satellite
spots.

2. If you can observe the signal pulses on your cytometer, watch the forward
scatter trace (any parameter will probably do).
Besides the usual superimposed pulses for most signals at the left hand edge
of the screen you will see a few pulses further to the right which occurred
within the sweep time of the oscilloscope - this is fine, but when a small
clump comes through and disrupts you will see a huge shower of pulses to the
right, some will probably be of excessive magnitude.

The solution is too filter or preferably disrupt these very small clumps.
Sometimes you can actually see, with the unaided eye, these fine clumps in
the sample tube as a sort of mist.

Hope this helps.

############################################################################

Email:	alan.bishop@imvs.sa.gov.au
Tel:	      +61 8 82223602
Fax:	      +61 8 82323139
Mail:	      Alan Bishop MA cantab.
	    Senior Flow Cytometrist
	      PO Box 14 Rundle Mall
	    Flow Cytometry Facility
	    Hanson Institue
	    IMVS
	      ADELAIDE SA 5000
	      AUSTRALIA



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:27 EST